Roche 454 GS FLX+ DNA Genome sequencer teardown

แชร์
ฝัง
  • เผยแพร่เมื่อ 10 พ.ย. 2024

ความคิดเห็น • 241

  • @davidallen6421
    @davidallen6421 8 ปีที่แล้ว +140

    Oh man a 454! I used to work on sequences generated on one of these!
    I took a few notes as I was watching:
    Filters/flow control on reagents: yes those are filters, either 0.33 or 0.22 micron. Pretty standard fare for mol-biol labs. They don't pass microbes, keeping your samples free of foreign DNA.
    Reagent cost: very high cost reagents is standard practice in this business, it's how they make their money. Now, you _can_ make your own (they don't tell you what the reagents actually are, but the reactions taking place in the machine are pretty standard DNA copy operations, so you can take an educated guess, then mass-spec/HPLC the real ones to see if you're right, and make up your own and give it a go). Having said that, the reaction chamber slide thingy (can't remember what it's called) is a consumable (ie can't be re-used, because how do you *know* you've cleaned all the DNA from the last run off and that it won't compromise the next one?) and it probably comes as part of a reagent kit for the machine, so they've got you over a barrel there.
    Peltier-cooled reagent reservior lack of active temp control: not necessary. The actual enzymatic reactions taking place within the machine are controlled, but all that peltier is doing is keeping reagents chilled so they don't denature or degrade at room temp during the course of the run (pretty common for biol stuff dealing with reactions producing light). The actual temp isn't really important, most labs just use domestic fridges/freezers for this kind of storage.
    Labour-intensive sample-prep: Yep, that's a given. To get the most out of this machine, you need to provide it with very carefully measured samples. The DNA to be sequenced needs to be sheared into lots of strands of a particular lenth, have special joiners and the beads attached then introduced at a very well defined concentration so you don't overload the sequencer. High-speed cetrifuges for seperating DNA are everywhere in mol-biol labs, so they expect you to have access to standard equipment. Basically, you pretty much expect to employ someone full time to look after the machine. They'd feed it samples, plus do basic analysis to get the crude data to something you can begin to analyse.
    The plate behind the reaction chamber: Should be held at 37 deg C for the DNA polymerase reaction.
    That camera is really interesting, I hope you can get it up and running. Thanks for another great video Mike :)

    • @cliveramsbotty6077
      @cliveramsbotty6077 8 ปีที่แล้ว +1

      Would you be so kind as to give a brief simple explanation of what the machine is used for? Thanks

    • @davidallen6421
      @davidallen6421 8 ปีที่แล้ว +5

      No probs. I've already done so a little further down the comments thread. th-cam.com/video/XaumUp4GpCw/w-d-xo.html&lc=z124jfo5nlepwjjis04cepwzgwanyjjowzc0k.1481463132881029
      That link was meant to take you there, but it doesn't seem to work. Sigh. Scroll down and look for a comment by TheAmmoniacal and you'll find it in the reply thread.

    • @matiastripaldi406
      @matiastripaldi406 5 ปีที่แล้ว

      Just a question. Why would you need a DNA polymerase reaction in a sequencer? Is it like a PCR reaction where you multiply the DNA fragments to make them easier to sequence?

    • @topoisomerase
      @topoisomerase 3 ปีที่แล้ว +1

      @@matiastripaldi406 most DNA sequencing is done by synthesis. By that I mean you produce the complementary strand of what you're sequencing, and the DNA bases (A,C,G,T) you add to make that strand either generate a product (like PPi with 454) or include a modified base (ILMN and most others) that allows you to detect that a DNA base was added. DNA polymerase is the enzyme that adds the bases to the growing complementary DNA strand. You do this sequentially and as such read the sequence. There is often a PCR step (or other DNA amplification technique) upstream of the sequencing reaction itself as well. That PCR may or may not use the same DNA polymerase as is used in the sequencing step.

    • @Tadesan
      @Tadesan 2 ปีที่แล้ว

      Your shortening of biology to “biol” gives me really uncomfortable feelings.
      How do you pronounce your abbreviation?

  • @douro20
    @douro20 8 ปีที่แล้ว +5

    Those heaters are from a company to the north of me in Leavenworth, KS. They are the world's largest manufacturer of thin-film heating elements and also one of the largest manufacturer of ceramic PCBs.

    • @MaxKoschuh
      @MaxKoschuh 8 ปีที่แล้ว

      ceramic PCBs, yummy :-)
      I would love to build a low noise audio preamplifier around a MIL-spec OP-AMP on a ceramic PCB. Would be neat.

  • @philorkill
    @philorkill 8 ปีที่แล้ว +4

    Thank you for this. I'm always impressed how fast this super complex pieces of kit can get obsolete.

  • @Mythricia1988
    @Mythricia1988 8 ปีที่แล้ว +22

    "Possibly a few astronomy type people wetting themselves" Damn right... Holy shit the pixel pitch. Big humongous (back-illuminated!!) buckets for light gathering. Like, holy crap. 7.4um is considered "rather large" as far as astronomy CCD's go. This one has 13.5um pixel pitch. Compare this to instruments from Santa Barbara Instrument Group (SBIG), and I think I'd safely value this camera pretty highly - you can easily pay upwards of 10'000$ for a 16mpix 7.4um cooled CCD camera, and this has larger pixels which is a pretty attractive feature. And it's cooled! So are many amateur CCDs, but the combination of massive huge pixels and very low-temp capability means the noise figures on this thing must be incredibly good under optimal conditions. Looking at the QE (Quantum efficiency, so, the ratio of incoming photons turned into readable electrons), this thing goes right up over the 90% mark at some wavelengths. For comparison, a typical color CMOS sensor used in a DSLR camera gets up to a whopping 35% or so.
    Damn. I don't care that I'd never be able to mount this on my 8" SC telescope (this camera is almost as large), nor power it, nor extract any images from it, but damn.

    • @Keith_Ward
      @Keith_Ward 8 ปีที่แล้ว +2

      Looks like the CCD on this model is 4k x 4k

    • @Mythricia1988
      @Mythricia1988 8 ปีที่แล้ว +2

      Yup, in the 16Mpix range, which combined with cooling and huge pixel pitch, and back-lit array, puts this camera comfortably in the "obscenely expensive" category if compared to off-the-shelf astronomy cameras. Heck it would be extremely good even in the 2k format... High resolution is not necessarily as important in astronomy - a 4k by 4k camera often ends up oversampling images on most telescopes. Meaning the optical resolution of the telescope is worse than the pixel resolution of the camera. Binning can still help make use of the excess resolution though, so if you can have high resolution AND big pixels, it's fine. Normally you have to pick one over the other.

    • @WickdPerfekT
      @WickdPerfekT 6 ปีที่แล้ว

      Mythricia mate, all you need is a larger mount bahaha

  • @Lactuca
    @Lactuca 8 ปีที่แล้ว +6

    "mixing small volumes of very expensive clear fluid" That's how I describe my job and that's really all it is.
    Done lots of 454 sequencing over the years, never had the chance to take one apart though. Done a few qRT-PCR machines though

  • @hene193
    @hene193 8 ปีที่แล้ว +23

    And when the wife asks what did you buy this time. "Oh just an old obsolete genome sequencer"
    I guess everyone has their hobbies.

  • @bborkzilla
    @bborkzilla 8 ปีที่แล้ว +32

    $500,000.00 to $250.00 in 9 years? That's extreme obsolescence.

    • @turboslag
      @turboslag 8 ปีที่แล้ว +2

      The pace and price of progress!

    • @grndkntrl
      @grndkntrl 8 ปีที่แล้ว

      The machine was manufactured in 2012. It's now 2016. That's only 4 years.

    • @turboslag
      @turboslag 8 ปีที่แล้ว +2

      kyoukoku
      That's an eternity in electronics and high tech! How many iphones and ipads have there been in that time?

    • @alerighi
      @alerighi 8 ปีที่แล้ว +1

      This must be declared illegal, it's pure planed obsolescence... also doing this for medical stuff, it's a lot worse, first at all, because the price is charged to the state (at least in counties with medical assistance, like all the European ones) and the money goes to a company that makes work and pay taxes in other countries, and also because the right to health it's a constitutional one, so i think it should be illegal that private companies make billions of billions at the expense of sick people and the state (and the citizen taxes)

    • @jcnossen
      @jcnossen 8 ปีที่แล้ว +10

      I don't think it's planned. The cost of the runs with all the expensive reagents (as he mentioned a run costs about 6000$) quite quickly make the machine obsolete, as there will be a new machine that's able to reduce the end user price of DNA sequencing. It would be against the patients interest if they keep using this machine too long.

  • @AdamWHankins
    @AdamWHankins 8 ปีที่แล้ว +1

    I also live right across the street from Spectral Instruments here in Tucson. good friend of mine works for them in development

  • @landtechnik4k
    @landtechnik4k ปีที่แล้ว +2

    when a UFO crashes, we all would like to watch mike disassemble it
    for the meantime we watch him undo human made crazy tech

  • @gglovato
    @gglovato 8 ปีที่แล้ว +14

    keep the UPS that's a top of the line unit!

    • @ChipGuy
      @ChipGuy 8 ปีที่แล้ว

      I think he has already given that away, according to his tweets...

    • @beyonddeath123
      @beyonddeath123 8 ปีที่แล้ว

      maybe several years ago, thats only worth 200-250$ us with good batteries

    • @gglovato
      @gglovato 8 ปีที่แล้ว +2

      JH
      UPS don't really depreciate, specially APC online ones, they last forever and you only need to change batteries there's no reason for them to depreciate, at least here in my market 10+ yr old UPS still cost the same as a new one

    • @beyonddeath123
      @beyonddeath123 8 ปีที่แล้ว +1

      I only know as I bought one recently, companies get rid of them after lease, so you can get them cheap if you know where to go. I bought a 3000va with new batteries for 350 cdn pesos, though moving it was a bitch.

    • @Mythricia1988
      @Mythricia1988 8 ปีที่แล้ว +2

      Would be worth quite a lot in some countries. Like here in Norway you pay 100$ for a piece of shit 600va unit made in China. You've no right to complain about your Canadian pesos just yet! :P

  • @RadRider33
    @RadRider33 8 ปีที่แล้ว +4

    Mike, in regards to the metric observation: I work at an American medical device company and all we do is metric. Much cleaner and simpler. Everyone groans when we need to use IPS...

  • @ahelms
    @ahelms 8 ปีที่แล้ว +3

    I work for a custom machine integrator company in the US and we only use metric. We only use US fasteners whenever forced by the customer.

  • @Konecny_M
    @Konecny_M 8 ปีที่แล้ว +2

    Aluminium can be readily nickel plated, that explains the look of the heatsink. Makes sense if any portion of it is exposed to high vacuum section (anodising is a no-go there due to the porosity, hence creating nasty virtual leak).

  • @MaxKoschuh
    @MaxKoschuh 8 ปีที่แล้ว +6

    Mike, this is wonderful. Thank you!

    • @ChipGuy
      @ChipGuy 8 ปีที่แล้ว +2

      Yeah I enjoyed that bit of kit very much. Thanks Mike. Somehow those valve blocks with the SMC valves on them are kinda sexy. I'd prefer the smaller Bürkert valves though.

  • @arcadeuk
    @arcadeuk 8 ปีที่แล้ว +6

    Wow, I love these cool biotech teardowns. That's Patreon money well spent!

  • @JacGoudsmit
    @JacGoudsmit 8 ปีที่แล้ว +23

    I can see how you misjudged the size of this thing. At the start of your video it really looks like it's maybe 24 inches (60cm) wide but at 7:06 your hands showed up and I went HOLY CRAP THAT THING IS BIG. :-)

    • @mikeselectricstuff
      @mikeselectricstuff  8 ปีที่แล้ว +5

      I did mean to put something in front for scale but forgot.

    • @JacGoudsmit
      @JacGoudsmit 8 ปีที่แล้ว +1

      Your hands do just nicely as a size reference, thanks :-)

    • @Tinker001
      @Tinker001 8 ปีที่แล้ว +1

      Until you realize he's actually a professional basketball player & his hands are twice the size of a normal persons...

  • @TheAmmoniacal
    @TheAmmoniacal 8 ปีที่แล้ว +18

    Been studying this stuff for many years as a biochemist, but still never got to use one :( Very interesting to see one tore down to compare theory with reality! Know exactly how it works.

    • @uski
      @uski 8 ปีที่แล้ว +11

      Dare to explain for non-biologists like us all ? :)

    • @GoldSrc_
      @GoldSrc_ 8 ปีที่แล้ว +1

      Yeah dude, I want to know how that thing works.

    • @davidallen6421
      @davidallen6421 8 ปีที่แล้ว +70

      It's hard to explain this in a youtube comment, but I will give it a go (note, this is way simplified):
      First you need to know something about DNA. It's double stranded, each strand is like a string of letters (called bases). There are four letters that make up the strands: A, T, G and C. The two strands are wound around each other in the famous double helix due to their complementary nature. Each letter has a complement, A complements T, and G complements C. So you might get a double strand like this (the 5' and 3' is mol. biol. jargon denoting the direction of the strands) :
      5'-ATGCCGTAGCAT-3'
      3'-TACGGCATCGTA-5'
      The complimentary bases of one string stick to those in the other string via hydrogen bonding, which is what holds the double helix together. It's an important property that is leveraged by the sequencer.
      Okay, so you do a whole genomic DNA extraction from your organism of choice. Your sample will contain millions of cells, which means you get millions of copies of the chromosomes. You take your purified chromosomal DNA and shear it up into lots of little pieces approx a few hundred bases long. Now you 'ligate' (attach) some very short pieces of DNA with a known specific sequence to the ends of all your unknown sheared-up chromosomal sample DNA (these are adapters). You then denature the double strands so they fall apart into single stranded pieces.
      Take the single stranded pieces and add them to a suspension containing little tiny manufactured balls with short pieces of single stranded DNA (ssDNA) complimentary to the special adapters you stuck on your sample previously. Your ssDNA will be able bind to the adapters on the balls via complementary pairing, forming short double-stranded DNA right near the ball, with the still-unknown rest of the single strand floating off it. The balls outnumber the ssDNA pieces many times over, so on average, you only get one ssDNA piece on a ball and most won't get any DNA at all.
      Now put these balls onto the 'nano-titre' flow cell slide (the glass thing with the strips in front of the camera in mike's video). This thing has tiny little nano-sized holes in it. The balls will fall into the holes, taking the ssDNA strands with them. Think of it kind of like a nano-sized series of honey comb cells. Now there are so many balls and so little DNA relative to the numbers, that on average, there should be only one ssDNA piece per honey cell (if there even is one in there at all). The reason they are in the holes is, we can now gently wash reagents over them and do reactions without the balls (and hence DNA) moving relative to the camera (which is very important as we'll see).
      Okay, now comes the important part. The actual reaction. The reaction is to take your ssDNA pieces and make a copy of them one base or letter at a time. The enzyme that does this is called DNA polymerase. DNA polymerase cannot start on a single stranded DNA strand, it MUST start from double stranded and extend the complimentary strand. It's a good thing we have a very short piece of known double stranded DNA formed from the complimentary binding of the adaptors sticking the ssDNA to the the ball, huh? ;)
      Confused? Yeah, me too. It's ASCII art time!
      5'-ATG(*DNA pol.*) -> enzyme moves along making a new strand complimentary to the original
      3'-TACGTACGATCATGCTAAG-5'
      3'-TACGTACGATCATGCTAAG-5'
      5'-ATGCAT(*DNA pol.*) ->
      3'-TACGTACGATCATGCTAAG-5'
      5'-ATGCATG(*DNA pol.*) ->
      3'-TACGTACGATCATGCTAAG-5'
      And so on and so forth...
      The trick here in this machine is that the reagents they use are fluorescent. So each letter has a different fluorescent dye attached to it. We'll also only let the reaction proceed for one step at a time. You add the enzyme and the reagent for one base (say A). All the ssDNA pieces in the honey comb cells that start with a T (complimentary to A) will have an A added to them by the enzyme. After the cycle, the dye is excited and induced to fluoresce and the camera takes a photo of the slide. Then, you move on to T. Again, if there are any strands that now have an A exposed next in line, the enzyme adds the T, and you take a photograph. Then G, photo; then C, photo. This cycle keeps on repeating through all the letters over and over until you have a complete copy of the original strands and a series of photographs taken along the way. The enzyme can only add one base at a time because you only give it one base at a time to add.
      In order to get the actual sequence, we simply (HA!) analyse the photographs. NASA star field algorithms are used to look for changes to the dots in the pictures as the cycle progresses. Each dot point represents one DNA strand in a nano-well. As the colour of the dot changes from one cycle to the next, the colour in the photo will change from one to the next. The algorithm simply reads off the sequence using the colour as a marker. At the end, we have the complimentary sequence to the ssDNA strand that was in each well. Then the hard part is putting all the pieces of the puzzle together (assembling the sequence) to get a complete unbroken sequence of letters from one end of the chromosome to the other.
      I strongly recommend you have a read of the wikipedia articles on DNA and DNA sequencing. There are some good animations on youtube of the old-school sequencing methods (still in use), like the Sanger method. If you understand how that works, you will see this is just a fancy pants way of doing it the old way :) It just does it massively parallel.

    • @kosikko4400
      @kosikko4400 8 ปีที่แล้ว +8

      Thanks a lot buddy for explaining this process, very interesting !

    • @davidallen6421
      @davidallen6421 8 ปีที่แล้ว +14

      You're welcome. I've learned so much from Mike and other knowledgeable people in the comments of his videos, that it's good to be able to return the favour and be _that guy_ that knows something about an obscure and expensive piece of equipment. Cheers :)

  • @msfraser82
    @msfraser82 8 ปีที่แล้ว

    The optical connector on the camera is an MT-RJ duplex multi-mode. You can get multimode patch leads which convert this to multimode ST connectors, which would connect to a standard gigabit ethernet interface (it said gigabit ethernet on the interface board silkscreen). I've only seen the MT-RJ connector used on some old fixed-configuration 100Mbit Cisco switches.

  • @vincei4252
    @vincei4252 7 ปีที่แล้ว +1

    21:40 Been tearing down one of these things but I actually managed to get one that is fully working with the computer and all the software. I couldn't get the camera cover open but Mike pointed out that there is button on the computer to open the cover in a video I sent him which I totally missed. Once pressed it takes quite a while and some whirring before the cover unlocks and you can open the camera.

  • @frollard
    @frollard 8 ปีที่แล้ว +1

    8:22 looks like the molding process where you have a negative mold, and you paint the polished smooth cavity (also acts as mold release), then either fiberglass or vacuum form or thermoset or resin cast the internals in a tumbling mold form. It's how they make boat hulls. (see the episode of how its made re: sea-doo watercraft)

  • @Sixta16
    @Sixta16 8 ปีที่แล้ว

    Nice to see new videos coming!

  • @davidallen6421
    @davidallen6421 8 ปีที่แล้ว +3

    I'm a little surprised to see this machine was manufactured so late (2012 iirc). By that point, there were far better 2nd gen sequencing options available that produced orders of magnitude (literally) more data and deeper sequence coverage for roughly the same cost. My department purchased an Illumina machine in about 2010 that blows the specs of that 454 out of the water. Similar upfront cost, but slightly more cost per run. However, it produced so much more data per run, that you could actually put multiple genomes on the slide at the same time (making sure to use a special DNA labelling technique to tell the final data apart) and do crude 'multiplexing'. This further reduced the cost/genome sequenced.
    Actually, the reason might be that although 454 didn't get as good depth of coverage, the legths of the DNA reads were much longer than more modern machines (100s vs 10s of bases), making it easier to actually assemble your sequence data into a genome. This is very useful if you're sequencing something that's never been sequenced before, but somewhat useless if you've already got a genome to compare to.

    • @Tadesan
      @Tadesan 2 ปีที่แล้ว

      So instead of getting more useful data upfront it gives you a whole lot more data that need a lot more processing to be meaningful?

  • @JBaughb
    @JBaughb 8 ปีที่แล้ว +3

    This was fantastic. I love teardowns of over engineered tech.

  • @palharley
    @palharley 8 ปีที่แล้ว

    The Detail of your videos is fantastic. I love when you connect to the SPI and dig into the the underbelly of the code.

  • @andycristea
    @andycristea 8 ปีที่แล้ว

    Awesome video! Thank you! Can't wait to see the camera video!

  • @campbellmorrison8540
    @campbellmorrison8540 2 ปีที่แล้ว

    The trolley is really nice. I have pulled a fair bit of lab stuff apart and Im pleased to see your hands are dirty too. Its surprising how dirty gear that is in a very good environment gets

  • @Tadesan
    @Tadesan 4 ปีที่แล้ว

    those cables just actuate the mechanism...
    that is beautiful

  • @stonent
    @stonent 8 ปีที่แล้ว

    Tubes marked W and R, W for water (the external port) and R for Reagent. So likely flush with reagent and then a final flush with distilled water.

  • @christopherleubner6633
    @christopherleubner6633 2 ปีที่แล้ว +2

    That camera has an image intensifier with a fiber taper attached to it. You could build a crazy sensitive astronomy or microscope camera with it. 🤓

  • @Darkassassin09
    @Darkassassin09 8 ปีที่แล้ว

    Defiantly fascinated to see what you can do with that camera when/if you get it running

  • @altimmons
    @altimmons 2 ปีที่แล้ว +1

    I’m glad to see metric hardware on it, it’s so much more sensible - anyone in the sciences should have an appreciation for the metric system because it’s a sane system

    • @altimmons
      @altimmons 2 ปีที่แล้ว

      I live in the US and in my technical work i typically work in metric

  • @SionynJones
    @SionynJones 8 ปีที่แล้ว +2

    Thanks Mike

  • @kd1s
    @kd1s 8 ปีที่แล้ว +1

    And I can tell you there's a missing component. There should be a server with heavy disk space behind the unit. Did a contract back in 2013-2014 doing the IT side of sequencing exomes. Massive amounts of storage and encrypted drives. They used Ion Proton sequencers.

    • @kd1s
      @kd1s 8 ปีที่แล้ว

      Oh and yes, reagents are kept very cold/frozen. And the process you'll find has a hell of a lot in common with developing color film.

  • @johnkeates9434
    @johnkeates9434 8 ปีที่แล้ว

    It is standard gigabit ethernet over optical fiber. You can use a standard fiber-to-copper adapter. The connector type and fiber type denotes a wavelength, which would be important to get right for the converter box ;-)

  • @altimmons
    @altimmons 2 ปีที่แล้ว

    The way this would work is that you would have mixed in with the bases ACTG some tagged defective ACTGs with fluorescent probes such that copies of dna would terminate at random points - this gives you a blast and you have to use the computer algorithm to line them back up- different length sequences will come off the column at different times based on charge vs drag and the camera would like read the fluorescent probe

  • @richardparry7532
    @richardparry7532 8 ปีที่แล้ว

    The fibre connector is an MT-RJ. It used to be popular about 10 years ago for Ethernet connections but replaced by LC which is still the standard. You can pick up MT-RJ network cards on eBay for a few pounds, although not sure what speed that runs at. The connection could be 10, 100 or maybe 1000Mbps. I'd also be surprised if it's Ethernet though, as unlikely there is IP connectivity going on in the camera. I'd hazard a guess it has some special card that communicates. Great video Mike.

  • @catt87
    @catt87 8 ปีที่แล้ว

    Really looking forward to you having a play with that camera module.

  • @bearb1asting
    @bearb1asting 8 ปีที่แล้ว

    Really dig your work, and your teardowns.

  • @sdshlanta
    @sdshlanta 8 ปีที่แล้ว +3

    +mikeselectricstuff as for why i is metric, the us medical industry only uses metric measurements.

  • @semidemiurge
    @semidemiurge 8 ปีที่แล้ว +12

    I live about 3 miles from Spectral Instruments in Tucson, AZ. Not sure if that helps at all.

    • @mikeselectricstuff
      @mikeselectricstuff  8 ปีที่แล้ว +47

      Do they have a dumpster?

    • @semidemiurge
      @semidemiurge 8 ปีที่แล้ว +8

      I will definitely check tomorrow. Also, a lot of the engineers and scientists in the area are avid cyclists and I think I might have ridden with a SI employee on a recent group ride.

    • @semidemiurge
      @semidemiurge 8 ปีที่แล้ว +5

      This telescope uses the 800 series.
      slotis.kpno.noao.edu/LOTIS/system.php

    • @MaxKoschuh
      @MaxKoschuh 8 ปีที่แล้ว +3

      such a small world

  • @blank7921
    @blank7921 8 ปีที่แล้ว

    To determine if it is molded look for the ejector pins. Or molding marks things like date codes etc will stick out like a sore thumb. It is possible it is urethane cast.

  • @DavidMills_Physicist
    @DavidMills_Physicist 8 ปีที่แล้ว +7

    I've got a board for the camera if you wish to borrow it for a few weeks.

    • @mikeselectricstuff
      @mikeselectricstuff  8 ปีที่แล้ว +2

      I found a few on ebay - only worth getting if I can get the software - do you have that ?

    • @DavidMills_Physicist
      @DavidMills_Physicist 8 ปีที่แล้ว +8

      Yep - got the software too. Sent you an email.

  • @TheDefpom
    @TheDefpom 8 ปีที่แล้ว +3

    They would still injection mould, just using aluminum mould instead of steel.

  • @the6p4c
    @the6p4c 8 ปีที่แล้ว

    Damn, I really wanted to see that camera working. Looking forward to that next video, if it comes.

  • @stevec00ps
    @stevec00ps 8 ปีที่แล้ว +2

    (Half way through so you may have already answered this) Not sure it's standard Ethernet fibre - "Fiber optic communication; standard communication to computer by fiber optic cable to proprietary PCI or PCIe card." on www.specinst.com/Brochures%20Rev%20B/800S-camera-broch_revB.pdf

  • @msylvain59
    @msylvain59 8 ปีที่แล้ว +2

    The construction seems to be a lot more clean and tidy than the Horiba blood analyzer.

  • @commonmogoreanu7135
    @commonmogoreanu7135 8 ปีที่แล้ว

    Great video! Thank you Mike!

  • @DextersTechLab
    @DextersTechLab 8 ปีที่แล้ว +4

    Nice score Mike, i was watching that one for a while and would have had it but i was knee deep in Quantel shite :-P Will watch with interest tomorrow.

  • @DoRC
    @DoRC 8 ปีที่แล้ว +1

    by the spectral instruments brochure you have yourself a 4k by 4k camera with a 27mm sq sensor. Dude that thing is incredible. I really hope you can get it to do something cool.

    • @mikeselectricstuff
      @mikeselectricstuff  8 ปีที่แล้ว

      Not necessarily - there are several CCD options listed

    • @DoRC
      @DoRC 8 ปีที่แล้ว

      they have pictures of the different models and yours matches the top of line one!

    • @DoRC
      @DoRC 8 ปีที่แล้ว

      mikeselectricstuff oh ok. the brochure made it look like the one that is in the form factor like yours is the 4k one.

  • @MrHack4never
    @MrHack4never 8 ปีที่แล้ว +22

    Hey Mike, jugding from all your medical equipment teardowns, are you planning on getting a MRI scanner for teardown?

    • @Cofcos
      @Cofcos 8 ปีที่แล้ว +9

      Having been in an MRI several times, I'm sure they must weight literal tons. And when removed from their site are already torn down into unrecognizable parts.
      Though, if he could manage to get invited to a teardown of an MRI machine, I'd love to see it...

    • @FrozenHaxor
      @FrozenHaxor 8 ปีที่แล้ว

      Lol.

    • @BenjaminEsposti
      @BenjaminEsposti 8 ปีที่แล้ว +1

      He should get one of those 100kV+ xray transformers! X3

    • @FrozenHaxor
      @FrozenHaxor 8 ปีที่แล้ว

      +Benjamin Esposti He already does.

    • @Electronichub_05
      @Electronichub_05 3 ปีที่แล้ว

      @@FrozenHaxor i need one too haha
      The arcs must be insane

  • @apathyisanartform
    @apathyisanartform 5 ปีที่แล้ว

    Pyrosequencer. DNA pattern adheres to the beads and and bases are introduced bound to ATP. ATP is then cleaved by luciferase (lightning bug enzyme) creating a small flash of light. The camera must capture that.

  • @heathwellsNZ
    @heathwellsNZ 8 ปีที่แล้ว

    Many thanks for such an interesting video as always.

  • @io504
    @io504 8 ปีที่แล้ว +3

    I very much approve the Pink Floyd server :) Fantastic video mike, really enjoyed the detailed teardown

  • @ejayerik
    @ejayerik 8 ปีที่แล้ว

    Incredible teardown.

  • @icesoft1
    @icesoft1 8 ปีที่แล้ว +1

    That fiber-optic connector looks to be a standard MTRJ connection, and if the camera was also offered in Gig-Ethernet, it's likely also Gig-E over a different medium (besides the usual copper). Being this recent, it probably is actually Ethernet, plugged into a matching SFP module plugged into a fiber Ethernet card on the PC/Server.

    • @mikeselectricstuff
      @mikeselectricstuff  8 ปีที่แล้ว +3

      Looked into it some more and it has Spartan 2 FPGAs so a pretty old design, now less optimistic that it will be Ethernet.

    • @icesoft1
      @icesoft1 8 ปีที่แล้ว +1

      Looking at some of the sales literature, you are probably correct. I turned up what looks to be some of the proper interface cards (perhaps), have a look on eBay for "Spectral Instruments camera interface card", there seemed to be a few of them on there that matched the interface description of 'Either 50-pin connector or fiber-optic' (per the sales literature).

  • @KingOfKYA
    @KingOfKYA 8 ปีที่แล้ว

    For fiber google "MTRJ (Male) to LC multimode " the orange means it will likely be 1gbit. (the lc connector will fit most pcs/switches) with a 850nm sfp, but it may be another odd feq. Good news is all of that fairly cheap second hand:P

  • @AureliusR
    @AureliusR 8 ปีที่แล้ว

    Mike, reading the PDF you linked shows the rack of tubes you have is specifically for the pre-wash stage. You would fill those tubes with the Pre-Wash Buffer after cleaning out the casette from the previous run. So that tray would not be in there during normal operation -- it does indeed come with cassettes (as they call them) which slide in the same way that tray does. Also, the back tube is chilled, as it contains the Sequencing Enzymes (whatever they are) and they have to be kept below -15C from what I understand.

    • @AureliusR
      @AureliusR 8 ปีที่แล้ว

      That tray you have is referred to as the Pre-Wash Cassette.

    • @AureliusR
      @AureliusR 8 ปีที่แล้ว

      These are the kinds of instructions you get:
      Wash the appropriate BDD using a soft bristle brush and Sparkleen solution. Rinse
      thoroughly with nanopure water and let the device air dry on a paper towel.
      Nanopure water?!
      Add 6.8 µl of Apyrase to the 20 ml supplemented Titanium Bead Buffer in the conical
      tube. Re-cap and label the conical tube “EB Wash”. Mix gently by inversion, and place
      the conical tube on ice.
      4. Add 30.6 µl of Apyrase solution to the bottle of supplemented Titanium Bead Buffer.
      Re-cap and label the bottle “BB2”. Mix gently by inversion, and place the bottle on
      ice.
      6.8ul? How do you even measure that much?
      Vortex the Enzyme Beads and the PPiase Beads and place them in a Magnetic
      Particle Concentrator (MPC) for at least 2 minutes for the beads to pellet. Invert the
      MPC several times and wait at least 2 minutes again. Carefully remove the
      supernatants and then remove the tubes from the MPC.
      Right....

    • @AureliusR
      @AureliusR 8 ปีที่แล้ว

      Yes, that PicoTiter plate is replaced every time. It actually gets submerged in one of the prep liquids until it's ready to be inserted, which I found intriguing.

    • @mikeselectricstuff
      @mikeselectricstuff  8 ปีที่แล้ว

      That cooler gets to about freezing point at room temp, but then the heatsink overtemp cutout trips so probably doesn't run quite that cold.

    • @andrewmorton9683
      @andrewmorton9683 8 ปีที่แล้ว

      It's easy to measure microlitre quantities with capillary tubes - look up "tlc spotting capillary". (TLC = thin layer chromatography in this context.)

  • @rkan2
    @rkan2 8 ปีที่แล้ว +1

    There are few US-ebay listings which show the IBM-1U server and one of them also shows it running Red Hat with Roche software and three Firefox windows, which seem to have been opened at start..: :P

  • @Mulletsrokkify
    @Mulletsrokkify 8 ปีที่แล้ว +2

    That's not a standard ethernet fibre connector Mike. The usual standard LC or the older SC and ST look very different. That's an MTRJ connector, probably quite easy to get an LC to MTRJ fibre somewhere

  • @redtails
    @redtails 8 ปีที่แล้ว

    I love your teardowns. In comparison to other teardown youtubers which I follow, you have by far the most interesting items and you also seem to possess the most knowledge even beyond electronics.
    (it's also such a waste for this machine to be replaced after just a couple of years, though I can sort of understand if the reagents are super expensive)

  • @agoodm
    @agoodm 8 ปีที่แล้ว

    Most APC UPS have the ability to step up or down the voltage to keep output at an optimal level without going to battery unnecesarilly.

  • @Spirit532
    @Spirit532 8 ปีที่แล้ว +1

    The image sensor is most likely behind a tapered FOP.
    The fiber-optic output is most likely just GigE, but I'm almost certain the actual protocol is nonstandard. CameraLink over fiber is extremely rare and stupidly expensive.

    • @mikeselectricstuff
      @mikeselectricstuff  8 ปีที่แล้ว +2

      It isn't tapered -it's a 60x60mm sensor! Fibre protocol is Cypress Hotlink over fibre. Video coming shortly

    • @Spirit532
      @Spirit532 8 ปีที่แล้ว

      mikeselectricstuff Whoa! That's a huge sensor, x-ray direct imaging sizes! Explains the price now.

    • @Mythricia1988
      @Mythricia1988 8 ปีที่แล้ว

      60x60mm? That's downright stupid. In like a hilarious, good sort of way. I don't even know of any standard telescope mounting options for such a sensor - it's larger than the "jumbo" sized 2" optical backs on most large amateur scopes, 11" aperture and above. Practically all smaller scopes are only going to have a ~38mm diameter usable imaging circle or so.

    • @Spirit532
      @Spirit532 8 ปีที่แล้ว +2

      @100531782681373837774 Huge silicon is used in direct x-ray imaging. The most high-end~($50-200k) models have stitched sensors of over 30x40cm(but usually TFT, so not quite silicon).

    • @Spirit532
      @Spirit532 8 ปีที่แล้ว +1

      mikeselectricstuff I just found the sensor used in your camera.
      www.sta-inc.net/sta4150/

  • @ChipGuy
    @ChipGuy 8 ปีที่แล้ว +4

    28:00: Do I see a black patch of magic smoke on that board where the fan cable is attached or is that just a shadow?

    • @ChipGuy
      @ChipGuy 8 ปีที่แล้ว

      I am afraid that the camera is actually toast and since it seems to be the most expensive part of the entire machine (worth $500.000 new) it's the reason this unit has been scrapped in the first place.

    • @icesoft1
      @icesoft1 8 ปีที่แล้ว +6

      Looks like a 'solder-plated' portion of the board there, and you're seeing the 'dark' glare off of it from deeper in the camera...

    • @JWH3
      @JWH3 8 ปีที่แล้ว +2

      How did you infer it's toast? A small burn on one part of the board doesn't mean toast, could be repairable.

    • @ChipGuy
      @ChipGuy 8 ปีที่แล้ว

      It was just an assumption. And yes even that damage is most likely repairable.

    • @ChipGuy
      @ChipGuy 8 ปีที่แล้ว

      Nope I did not. It was a reflection since there is no soldermask on the board. Yay!

  • @DavidAlanGilbert
    @DavidAlanGilbert 8 ปีที่แล้ว +1

    +mikeselectricstuff The problem with pacemakers is that magnets are used for medics to switch them in the field; so it's not an issue with interfering with the operation so much as they're intended to switch mode; www.medscape.com/viewarticle/749751_5 (so for example pacemaker users are told not to put headphones in their shirt pockets).

  • @johnfranks
    @johnfranks 8 ปีที่แล้ว +3

    AVE needs one of these to finish his bartending robot.

  • @RhythmGamer
    @RhythmGamer 8 ปีที่แล้ว +2

    That fiber connector is MTRJ, not really that common but its for ethernet comm

  • @OneBiOzZ
    @OneBiOzZ 8 ปีที่แล้ว

    i have to say it looks like no fibre ethernet i have seen before, but i have seen them use all sorts of connectors internally inside fibre switches to route internally including just off the shelf SPDIF fibre so who knows
    it could be just a high speed serial interface like LVDS but you might want to try looking inside to see what chip is driving it and if there is a space for ethernet magnetics you might be able to solder on (such as gb or 10gb)

  • @ArtesianFalma
    @ArtesianFalma 8 ปีที่แล้ว +5

    It looks small to begin with, then you realise it's the size of a plane

    • @MaxKoschuh
      @MaxKoschuh 8 ปีที่แล้ว +1

      exactly my thoughts. I was surprised when I saw the first size comparison with Mike's hand.

    • @JohnWatkinsUK
      @JohnWatkinsUK 8 ปีที่แล้ว +2

      With the large sleeve, I wondered if mike had shrunk.

  • @stonent
    @stonent 8 ปีที่แล้ว +9

    Looks like this could be the camera controller: www.ebay.com/itm/Spectral-Instruments-PCBA-SCH-FO-Gigabit-I-O-3097-Rev-G-Camera-Interface-Card-/291956437009?hash=item43f9f5b011:g:kXsAAOSwo4pYFJgf It uses 850nm fiber, so it may be possible to use a standard 850nm fiber to ethernet media converter box to get into the camera. The camera brochure indicates an SDK is available for the camera upon request.

  • @joe72205
    @joe72205 8 ปีที่แล้ว

    Pacemakers (at one time) used a reed swtich to change operating mode (e.g. to high-activity). I don't think the magnetic field affects the actual operation of the device.

  • @proluxelectronics7419
    @proluxelectronics7419 8 ปีที่แล้ว

    Looks like Ethernet connection running "SI Image software suite" for config,

  • @bluephreakr
    @bluephreakr 7 ปีที่แล้ว

    50:55 was worth the wait.

  • @DoRC
    @DoRC 8 ปีที่แล้ว

    what kind of search terms do you use to to find things like this on eBay?

  • @unlokia
    @unlokia 7 ปีที่แล้ว

    Superb.

  • @NanoCottage
    @NanoCottage 8 ปีที่แล้ว

    Yes that fibre optic connector is standard, it's an MTRJ. They were very common on switches/routers around 15 years ago, hardly ever used today. A cable to adapt this to a more common format such as LC/SC/ST is easily sourced. I'd expect it to operate at 850nm, possibly stamped on the transciever in the camera? Can you see any visible light from the fibre connector when the camera is on? Apparently the interface spec is "AIA Camera Link" edt.com/downloads/pcidv_family-3.pdf Some frame grabbers here edt.com/product/pci-dv-fox/

  • @TheCaitlinlopez
    @TheCaitlinlopez 7 ปีที่แล้ว

    Where do you get this? the gel and electricity that this use cooling and heating are very expensive

  • @tcdontplay4914
    @tcdontplay4914 8 ปีที่แล้ว

    good stuff

  • @you-dont-know-me
    @you-dont-know-me 4 ปีที่แล้ว

    How do you obtain data from the serial port? I want to learn that!

    • @AureliusR
      @AureliusR 3 ปีที่แล้ว

      Dunno if you've learned in the 6 months since -- but it's pretty simple.
      RS-232 (aka serial) has a very distinct pattern. Whenever you see a 4-pin header near a processor, or near the edge of the board, there's a decent chance it'll be serial. RS-232 voltage levels are +/-15V, but often microcontrollers send the same signals with TTL levels (0/5V or 0/3.3V) instead. In either case, you just get a serial-to-USB dongle of some kind (there's about a billion different varieties floating around on eBay, or you can shell out for a nicer one from a supplier -- Adafruit have a good one) and connect it up. And voila, you'll get serial comms.
      Sometimes it'll be actually connected to a DE-9 port, in which case it probably *is* RS-232, and in that case you can (you guessed it) get an RS-232 to USB converter. The only difference between the two types of dongle is the RS-232 type has a MAX232 or similar chip that converts between the two voltage signalling levels.
      Let me know if you have any other questions.

  • @NickAlexeev
    @NickAlexeev 4 ปีที่แล้ว

    The large plastic parts could have been made by urethane casting. Then polished after they have been removed from molds.

  • @90SecondsofAviation
    @90SecondsofAviation 8 ปีที่แล้ว

    Mike, might we know how You find these things in ebay ? specific categories or what ?

    • @markhodgson2348
      @markhodgson2348 3 ปีที่แล้ว

      Computer waste or wee recycling ♻️

    • @90SecondsofAviation
      @90SecondsofAviation 3 ปีที่แล้ว

      @@markhodgson2348 yo man that was 4 years ago

  • @TheModerGuy
    @TheModerGuy 7 ปีที่แล้ว

    39:34 chime triggers an epileptic episode in mike

  • @mbainrot
    @mbainrot 8 ปีที่แล้ว

    the cooler does have a temperature sensor on it

    • @nRADRUS
      @nRADRUS 8 ปีที่แล้ว

      but another side of radiator ?

    • @tbbw
      @tbbw 8 ปีที่แล้ว

      They could be smart enuff to have figured out that if they feed X ammount of Amps into those peltiers X ammount of units of heat gets moved.
      Unless the role of that thing is to just keep the liquid frozen when it's not used and they could more or less run them on almost full blast until it was time to unfreeze it.

  • @davemould4638
    @davemould4638 2 ปีที่แล้ว

    Low production run injection moulds can be made pretty cheaply these days. There are several companies offering a complete service whereby you upload a CAD file of the intended finished plastic part to them and they make aluminium A and B moulds using a CNC machine fed essentially from the CAD file with minimum human interaction needed, and then they can produce a few thousand parts. Your design is analysed by their computer and you get feedaback a day or less after uploading your design to highlight any errors that would make it unmanufacturable such as insufficient draft angles for the chosen plastic or features that are too narrow & deep to machine, and also warnings about likely sink dimples or flow lines etc. An engineer will also discuss the positions they plan to place the gates and parting lines to ensure that it will not cause any problems. I used a company called "Protolabs" in the UK but there are many others.

  • @Soundfactory24
    @Soundfactory24 8 ปีที่แล้ว

    I still remember the time 40 Years ago when I took a mechanical kitchen alarm, which is, of course, much more complicated than this cheap device ;-)

  • @KeenanTims
    @KeenanTims 8 ปีที่แล้ว

    Definitely not a standard connector used in fibre Ethernet applications. It looks like an 'MT-RJ' style connector on the camera side, not common in Ethernet but could be. This pdf: www.specinst.com/Products/800s_brochure.pdf mentions a 'proprietary PCI card'.

  • @liamharrison3
    @liamharrison3 8 ปีที่แล้ว +1

    its weird, you dont get an idea of the size until he talks about the surface finish

  • @DatBlueHusky
    @DatBlueHusky 8 ปีที่แล้ว +3

    50:52 "ribbed for pleasure" lol

  • @JGunlimited
    @JGunlimited 6 ปีที่แล้ว

    Checked on Ebay. I think your videos bumped the price quite a bit ;-)

  • @JasperJanssen
    @JasperJanssen 8 ปีที่แล้ว

    The s/n on the camera is 1119, which rather puts an upper bound in the number of DNA sequencers made.

    • @docdaneeka3424
      @docdaneeka3424 8 ปีที่แล้ว

      Lower bound, surely.

    • @JasperJanssen
      @JasperJanssen 8 ปีที่แล้ว +1

      No. The serial number would probably be numbered through for all the other variants of the camera that didn't go into DNA sequencers. Obviously it doesn't tell you how many were made after this one, but it does tell you that no more than that many were made up to 2012.

    • @docdaneeka3424
      @docdaneeka3424 8 ปีที่แล้ว +1

      Then by that reasoning it doesn't tell you anything about how many DNA sequencers were made. I don't see how seeing serial number x tells you there are no other machines out there with serial number greater than x? I could go on ebay and buy the same machine with serial number 1120? How do you get that '1119' was the last one made in 2012?

  • @lorenzo42p
    @lorenzo42p 8 ปีที่แล้ว

    could you sequence the dna of the star child skull?! love your vids. we might never get to see these things if it weren't for you.

  • @pepper669
    @pepper669 8 ปีที่แล้ว +2

    Hope there wasn't any nasty stuff remaining in there.

  • @jayc2469
    @jayc2469 2 ปีที่แล้ว

    The case _sounds_ like GRP

  • @monelfunkawitz3966
    @monelfunkawitz3966 7 ปีที่แล้ว

    I think I went to high school with Genome Sequencer.

  • @ChipGuy
    @ChipGuy 8 ปีที่แล้ว +5

    The housing looks like a vaccum low pressure mould. Something like this www.citim.de/de/Polyamidguss-Vakuumguss

    • @MaxKoschuh
      @MaxKoschuh 8 ปีที่แล้ว

      very interesting, never heared of that before

    • @ChipGuy
      @ChipGuy 8 ปีที่แล้ว +2

      There is a smaller company that also does room temperature vaccum mould (Raumtemperatur Vakuumguss) into silicone forms: www.mfd-modelle.de/ They can include premade metal brackets, brass thread inserts and other stuff into the mold. Here is one example: www.ibpmt.com/hdc_e/adoc.asp?hdc_e/hdc_overview.htm This housing is produced with that vaccum mould technology

    • @Morbuto
      @Morbuto 8 ปีที่แล้ว

      Could be, but it's pretty expensive for a piece that big, especially since only a few companies have the vacuum chambers large enough for parts like this. Another option is RIM (reaction injection molding).

    • @Graham_Langley
      @Graham_Langley 8 ปีที่แล้ว

      +Morbuto RIM/PU was my thinking too.

  • @Fan119
    @Fan119 8 ปีที่แล้ว +2

    378 views and someone has already disliked it, they are subscribed and get notifications on new videos just to dislike.

  • @marcus_w0
    @marcus_w0 7 ปีที่แล้ว

    I'd kill to get one of these for under a thousand bucks! What do I have to search? "DNA Sequencer" didn't return any hits in ebay ;)

    • @mikeselectricstuff
      @mikeselectricstuff  7 ปีที่แล้ว +1

      Try Genome sequencer or Roche 454. Plenty in USA atm but Ebay search omits international search results from sellers who ask international bidders to contact them for shipping.
      Search via geo-ship.com to bypass this filter.
      Shipping cost will be huge unless you can get them to remove the UPS and frame, but make sure you get the PC (or at least the PCI card) and camera PSU.

  • @sajukkhar
    @sajukkhar 8 ปีที่แล้ว +1

    What's Mike's wife opinion on everything he buys and bring home?
    Joking. ;)

  • @stonent
    @stonent 8 ปีที่แล้ว

    Possible camera brochure: www.photonlines.co.uk/downloads/800s-camera-broch_reva.pdf
    "Back illuminated CCDs along with industry-leading low read noise creates the perfect camera for low-light level applications such as astronomy, bioluminescence and plate reading for drug discovery."

  • @mrfrog8502
    @mrfrog8502 8 ปีที่แล้ว +2

    01:00 BioTech squishy since stuff 😂

  • @alext9067
    @alext9067 8 ปีที่แล้ว

    Is that a Frankenstein machine?